The bandpass of a spectrophotometer is 10 nm. If a setting is at 540 nm, the wavelengths permitted to impinge on the sample fall within which range?

Enhance your knowledge with the Ciulla Clinical Chemistry Test. Study with interactive quizzes, featuring flashcards and multiple-choice questions. Each question offers hints and explanations for better understanding. Prepare effectively and excel in your exam!

Multiple Choice

The bandpass of a spectrophotometer is 10 nm. If a setting is at 540 nm, the wavelengths permitted to impinge on the sample fall within which range?

Explanation:
Bandpass defines the width of wavelengths that pass to the sample around the chosen center wavelength. With a bandpass of 10 nm and the setting at 540 nm, the transmitted light spans 540 nm minus 5 nm to 540 nm plus 5 nm, i.e., 535–545 nm. This symmetric range around the setting matches the 10 nm width. The other options either shift the center or enlarge the range beyond the stated bandpass: 530–540 nm would center at 535 nm; 530–550 nm would be 20 nm wide; 540–550 nm would center at 545 nm rather than at 540 nm.

Bandpass defines the width of wavelengths that pass to the sample around the chosen center wavelength. With a bandpass of 10 nm and the setting at 540 nm, the transmitted light spans 540 nm minus 5 nm to 540 nm plus 5 nm, i.e., 535–545 nm. This symmetric range around the setting matches the 10 nm width. The other options either shift the center or enlarge the range beyond the stated bandpass: 530–540 nm would center at 535 nm; 530–550 nm would be 20 nm wide; 540–550 nm would center at 545 nm rather than at 540 nm.